[Spm] Re: Spm Digest, Vol 62, Issue 2: Cleaning of TGT-01 grating
Bert Lägel
laegel at physik.uni-kl.de
Wed Sep 2 08:54:06 EDT 2009
Hi Chris,
Blow off the dust with a nitrogen gun. Then I would first try the standard
cleaning procedure, i.e. cleaning with acetone-IPA-DI water in an ultrasonic
bath. The acetone probably dissolves the glue/epoxy that holds the Si
grating on the metal disc - so you might want to get some epoxy to glue it
back on.
If that is not enough try 1-5min in Pirana (H2SO4: H2O2, 1:3). You will have
to remove the Si-chip from the metal holder for this! (Don't forget to rinse
plenty in DI water after Pirana- I usually do that in an US bath, change
water after 5min, 5x). Please check up on the Pirana procedure if you
haven't done this before!
Best to use cleanroom equipment/environment for cleaning.
Good luck!
-------------------------------------------------