[Spm] Tapping in Liquid - Phase vs. Amplitude images
allard katan
katan at Physics.LeidenUniv.nl
Tue Apr 15 18:41:42 EDT 2008
Hi Don,
As to your first question: Taking into account most modern AFM systems allow
the acquisition of many channels at a time, and hard disks are far below
1$/GB, I'd consider it silly not to record the phase when you have it
available. On older systems (like the Nanoscope IIIa's in our lab) you will
have to make a choice. I don't see why the arguments that go into this
choice should be different for air or liquid. If you feel phase imaging is a
useful technique in air, I am curious to hear why you are uncertain it will
be useful in liquid. Of course, not all samples will show relevant
differences in tip-sample interaction, but if they are present it is more
likely to show up in the phase than in the amplitude, assuming you have a
well-tuned feedback. Under that same assumption the amplitude image will
only show features very similar to the derivative of the topography in the
scanning direction. In the case of feature sizes exceeding the dynamic range
of the topography image (very small structure on top of big features) the
amplitude image may contain additional information not present in the
topography. In any case amplitude variations will show up in the phase
image, too. It is possible to deconvolve this to some extent, but it can be
difficult.
As to the second question: a good way to find the real resonance frequency
is to get an AFM with some type of direct-drive excitation, such as magnetic
excitation. Using the Brownian motion spectrum of your cantilever is just as
good (at least to find the resonance). Some AFM software allows you to do a
thermal tune, when not imaging, and there's some workarounds for others. But
I can recommend having a spectrum analyzer (we use a Stanford Research
770FFT, but many more options are possible) connected to your photodiode
difference signal all the time during your measurement. Once you've learnt
how to read it, it's amazing how much you can tell about what is happening
to your tip from the oscillation spectrum. Be aware that if you want to see
the thermal spectrum in liquid, using the stiff tips usually used for
tapping in air puts demands on your optical lever detection system that are
not often met by commercially available AFMs. We usually use either AC240
(~2N/m, 30 kHz in liquid) or Biolever (~0.02 N/m, 8 kHz) tips from Olympus,
or the 'F' triangular lever on the Veeco Microlevers (0.5 N/m, 35 kHz) for
tapping in liquid.
Tuning your drive just below the resonance is useful because of the
amplitude behavior, not so much because of the phase. In theory, tuning
slightly below resonance should set you on the net repulsive part of the
tip-sample interaction, tuning slightly above on the net attractive part.
Which of these you'd prefer depends on your goals, but what happens in
practice may depend on strongly on the details of settings and instrument.
In any case, the low quality factor in liquid and the large influence of
squeeze damping make this less of an issue. In contrast to most 'autotuning'
software I do not think that choosing a frequency that coincides with a
resonance of the liquid cell is very useful unless the driving amplitude
demands it.
Hope this helps,
Allard Katan
Post-doctoral researcher
Bio-AFM lab/Interface Physics group
Kamerlingh Onnes Laboratory
Leiden University
_____
From: Don Chernoff at ASM [mailto:donc at asmicro.com]
Sent: Friday, April 11, 2008 13:24
To: SPM List
Subject: [Spm] Tapping in Liquid - Phase vs. Amplitude images
My questions today concern whether phase imaging in liquid is useful and how
to set it up.
1) Is phase imaging in liquid a useful technique, as it is in air? Does it
have significant advantages over amplitude imaging?
As background, consider this:
A 1997 paper from Helen Hansma and colleagues reports an impressive set of
observations using tapping mode phase imaging to visualize DNA molecules in
liquid. [see
http://www.pubmedcentral.nih.gov/picrender.fcgi?artid=147043
<http://www.pubmedcentral.nih.gov/picrender.fcgi?artid=147043&blobtype=pdf>
&blobtype=pdf]
They reported:
Phase imaging is optimal at lower imaging forces than height
imaging (Fig. 3). This is a significant advantage for phase
imaging over height imaging, since there is less damage to
biological samples at lower imaging forces. In addition, phase
images remained clear at a 3- to 6-fold faster scan rate than height
images, 27.5 Hz in these experiments (Fig. 2), corresponding to
a tip speed of 110 mm/s. With faster scan rates images can be
captured more quickly for observing processes in the AFM.
...
We have not experimented
systematically with the amplitude images of moving DNA, but
those who can do amplitude imaging ... might find that
amplitude imaging is also useful for imaging moving
DNA at faster scan speeds and lower forces.
===
With the perspective of almost 11 years since that publication, what is the
situation of phase vs amplitude imaging in liquid?
2) When doing phase imaging in air, it is easy to find the fundamental
frequency of the cantilever and (we think) it is important to set the drive
frequency a little bit lower than the resonant frequency. In liquid, it can
be hard to choose the correct resonance. Assuming you can choose it, is
there a benefit to adjusting the drive frequency slightly in order to make
the phase response more reproducible? Are there any other hints the
experienced users can give?
regards,
Don
=============================================
Don Chernoff, Ph.D., President
Advanced Surface Microscopy, Inc. E-Mail: donc at asmicro.com
3250 N. Post Rd., Ste. 120 Voice: 317-895-5630
INDIANAPOLIS IN 46226 USA Toll free: 800-374-8557 (in USA & Canada)
web: http://www.asmicro.com Fax: 317-895-5652
[business activities: analytical services in AFM, AFM probes, consulting,
training,
calibration and test specimens, calibration and measurement software,
used NanoScope equipment.]
=============================================
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://spmlist.di.com/pipermail/spm/attachments/20080416/d428b4f8/attachment-0002.html
More information about the Spm
mailing list